Atlas scientific k1.0 calibration issues
Hello, I have the k1.0 ec probe, when I run through calibration process I’m not getting readings anywhere near what’s printed on the cal solution bottles. I switched the protocol on the ezo boards and I believe the default probe setting is for the k1.0 so I never switched that. What’s possibly going wrong?
@vdelgatto
This is typical for the EC probe. The readings will not respond accurately until you complete the entire EC calibration.
The pH probe does respond with the correct readings after each command.
You can send sensor78-1 cal,? and it will display the current calibration points
For the EC it is sensor78-2 cal,? if you have the hydro kit with pH, EC and RTD
This is typical for the EC probe. The readings will not respond accurately until you complete the entire EC calibration.
The pH probe does respond with the correct readings after each command.
You can send sensor78-1 cal,? and it will display the current calibration points
For the EC it is sensor78-2 cal,? if you have the hydro kit with pH, EC and RTD