Atlas scientific k1.0 calibration issues

Home Assistant automation projects, questions, etc. go here.
Post Reply
vdelgatto
LED-Curious
LED-Curious
Reactions:
Posts: 8
Joined: Fri Aug 13, 2021 9:33 pm

Hello, I have the k1.0 ec probe, when I run through calibration process I’m not getting readings anywhere near what’s printed on the cal solution bottles. I switched the protocol on the ezo boards and I believe the default probe setting is for the k1.0 so I never switched that. What’s possibly going wrong?
jaysal
LED Enthusiast
LED Enthusiast
Reactions:
Posts: 75
Joined: Sat May 30, 2020 7:49 am

@vdelgatto
This is typical for the EC probe. The readings will not respond accurately until you complete the entire EC calibration.

The pH probe does respond with the correct readings after each command.

You can send sensor78-1 cal,? and it will display the current calibration points
For the EC it is sensor78-2 cal,? if you have the hydro kit with pH, EC and RTD
stud
LED Lover
LED Lover
Reactions:
Posts: 143
Joined: Wed Jun 03, 2020 4:19 pm
Location: Canada

Nice @jaysal love your incentive of helping the next .
@vdelgatto You're on Tasmota ? And tasmota sees the device ( command: i2cscan ) and you're getting readings ?